Temperature sensor:
I used 5 pieces of glass diodes (1N4148) wired in series. A diode’s forward voltage has a temperature coefficient of -2.1mv/˚C. What that means is that the diode’s forward voltage (typically 0.6v at room temp) DECREASES as the temperature is increased. Having 5 of these diodes in series gives me a temperature coefficient of -10.5mV/˚C. That is a big change in voltage and it makes the circuit very easy. And these diodes are very cheap. Plus, the glass package makes it ok for use in alcohol vapors.
After soldering, the diodes are wrapped with PTFE tape for insulation then placed inside a length of 5/32” copper tubing with one end sealed.
Cable:
You may have some unused LAN cable lying around somewhere. This has 5 pairs of twisted wires inside and it makes a very good cable for your probes.
Voltage Reference:
I used LM336-5.0 voltage reference and as the name suggests it is 5.0v. This reference is VERY stable which means the thermometer will also be very stable. Once calibrated you can expect it to hold it’s calibration for more than a year. DO NOT use an ordinary zener diode in place of this reference because your readings will drift over time and ambient temperature change. Your thermometer is only as good as your reference.
Circuit Description:
The Temp sensor and voltage reference plus the trimming resistors form what is known as a “bridge” circuit. The op-amps (TL-082) serves to present a very high impedance to the bridge so as not to upset its balance. The diode sensor when forward biased will have a voltage drop of approximately 3v. The trimmer resistor is adjusted so that at 100˚C we get a full scale reading on the meter. The op-amp also provides gain, in this case 20x. The gain will multiply the tempco of the sensor to:
10.5mV/˚C x 20 = 210mV/˚C
And it is this voltage that we feed to the meter to get a reading.
Meter:
I am using a 0~100uA analog meter. It has an internal resistance of 1.9k ohms. (Do a search to see how you can measure the internal resistance of your meter). I wanted it to read from 75˚C to 100˚C because this is the range of temperature that is of interest to distilling. Above and below that temperature range is of no use to me so there’s no need to display it.
I will write on the meter’s dial “75....80....85...90...95....100” so I can read the temperature right off the scale on the dial. Since the sensor’s tempco has been gained by 20x we get 210mV/˚C. The difference between 75˚C and 100˚C is 25˚C. So that means the voltage across the meter will be 5.25v at full scale reading. To get 100uA (my meter) out of 5 volts means a resistance of
5v / 100uA = 50kΩ
Since my meter has an internal resistance of 1.9kΩ I need to put a
50kΩ - 1.9kkΩ = 48.1kΩ
Resistor in series with the meter. It can be seen on the schematic diagram. If you meter has a different internal resistance then you must compute for what series resistor you need to put in there so that the total resistance of the meter and series resistor is equal to 50kΩ.
12v Power Supply:
I use a “wall wart” power supply which you can buy from computer shops. You can also use a small 12v motorcycle battery or even a 12v car battery. Power consumption of the circuit is very low, less than 1 watt.
Calibration:
Immerse the probe in boiling water and turn the trimmer so that you get a full scale reading. Once you take the probe out of the water you can see that the meter reading comes down also. The component’s values were selected so as to get a 75˚C to 100˚C reading on the meter.
If you have another water bath and can hold the temp to a constant 75˚C you can alternately immerse the probe in each bath and adjust for gain and offset. You now have a VERY accurate thermometer (You need to have the gain adjustable in order to do this). I did not bother to do this extra step.
(to be continued)